During transport,an infant who is using a Bio-Med MVP-10 unit in the intermittent mandatory ventilation mode with 60% fractional inspired oxygen requires 3 L/min of oxygen and 2 L/min of air to maintain the fractional inspired oxygen and desired peak pressure.An E cylinder with 2200 pounds per square inch gauge of oxygen would last _____ minutes.
A) 88.8
B) 155.5
C) 207.3
D) 314.3
Correct Answer:
Verified