A satisfactory lateral lumbar spine radiograph is taken using 200 mA,0.6 sec,94 kVp,and 40 inches SID. If an increase in SID to 60 inches is desired,which of the following technique changes will produce a radiographic density most similar to the original?
A) 200 mA, 0.25 sec, 94 kVp
B) 200 mA, 1.33 sec, 94 kVp
C) 400 mA, 1.33 sec, 94 kVp
D) 200 mA, 1.33 sec, 108 kVp
Correct Answer: