You have just taken a CR radiograph of a patient's knee and the S# is 225.Unfortunately, the image has some degree of motion distortion and needs to be repeated using the same exposure settings.In repeating the image, the new image shows no motion but the S# is now 400.No changes were made to collimation, patient position, exposure techniques, etc.Based upon this new S# value,
A) exposure reproducibility is suspect.
B) the mR/mAs is equal.
C) the exposure reproducibility is within tolerable limits.