A manual stat protein analysis is performed on a patient sample.The patient's unknown sample had an absorbance reading of 0.6.The protein calibrator with an absorbance reading closest to the unknown's absorbance had a concentration of 6 g/dL and an absorbance reading of 0.5.Applying Beer law,calculate the unknown protein concentration in the patient sample.
A) 6 g/dL
B) 7.2 g/dL
C) 8 g/dL
D) 5.5 g/dL
Correct Answer:
Verified